Hello, QjuHy!
I am user of QHY11 #0101608. Camera was supplied by QHY dealer in Russia.
Camera test at room temperature was performed well.
After that camera was assembled with real set up of the observatory.
Ambient temperature -12C.
Thermoelectric Cooler was switched off.
Normal rate of readout.
Used programm - EZCAP.
I have investigated picture of defocused star in prime focus at different temperatures.
Temperature was measured in three points:
1) Reading from EZCAP main window (sensor temperature)
2) Digital temperature gauge of gas flow at the output of camera "grid"
3) Ambient gas temperature
Temperature 3) was about constant -12C
Temperature 1) was changed by external gas flow heater directed to camera air inlet with step 2C from -12 to +5C.
At switched off thermoelectric Cooler the temperature 1) was equal to 2) with deviation + /- 1C
Following effect is observed:
a) Star picture disappear if temperature 1) is lesser than -3C
b) Active sensor area (where signal takes place) decreased proportionally with temperature decreasing from +2 to -1C
The zip file with pictures at different temperatures you can load from place
http://mdeminsky.narod.ru/TEST.zipAbbreviation of file's name:
30s_-12C-12O.jpg means exposition 30s, temperature 1) equals -12C, ambient temperature -12C
Used software
EZCAP 3.11
QHYCCD QHY11 ASCOM Camera Driver 1.1.0
QHY11 3.0 Driver
As I know, it is second QHY11 camera with similar problem here in Russia.
I would like hear any comments / advices from you.
Thank you.
Maxim